This question has been sent to RenalWEB by a Nurse-Manager:
We have been going round and round with Baxter Healthcare over this medical device alert. Our position is that the problem is inside the machine and installing a GFCI will not solve their problem and in fact |
|||
|
GFI outlets are required in all wet locations. This is part of most state electrical codes. The potential for treatment interruptions is by far out-weighed by the safety of the staff and patient. GFIs not only protect equipment but personnel as well. You can check with your local electrician for the state regulations regarding GFIs in your area. Hope this helps.
[This message has been edited by Bioman (edited 05-28-2004).] |
||||
|
I'm shocked that they do NOT recommend using GFI outlets. As the others stated, it is MANDATORY in every state I've worked in. I personally would be more concerned with the safety of the patient and staff than an 'interrupted treatment'. It will only interrupt when it has to.
[This message has been edited by GTSCSS (edited 06-01-2004).] |
||||
|