<Kman>
Posted
I have a 2008K machine with 2000 hrs on it. Our TCD is 13.7, I like for my machines to be at the least 13.6, though I know they can be lower. The machine is stuck at 13.5 and will not go any higher. It also takes a while to stablize at this conductivity. Any ideas? Thanks!
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
check and/or calibrate your acid and bicarb pumps, then calibrate your conductivity and see what happens. When you have your condo. meter hooked up does it read the same as your display?
 
Posts: 88 | Registered: 16 September 2009Reply With QuoteEdit or Delete MessageReport This Post
<Kman>
Posted
Joebiomed,

Just re-calibrated acid/bic pumps. Went to do conductivity, and now it wont go above 13.3!
 
Reply With QuoteEdit or Delete MessageReport This Post
<Kman>
Posted
Oh, ya the external matches machine.
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
Ok, check the filters in the concentrate connectors and look for any obstructions in the pumps flow path. Also try to determine if it's the acid or bicarb by just running with the acid and your condo should get to right about 10.0. What I would do after that is start to follow the troubleshooting manual for conductivity problems. Good luck
 
Posts: 88 | Registered: 16 September 2009Reply With QuoteEdit or Delete MessageReport This Post
<Kman>
Posted
ha, fmc troubleshooting book? PLEASE, that has hardly ever helped me before, you never find the problem using their stuff
 
Reply With QuoteEdit or Delete MessageReport This Post
<Kman>
Posted
Like I said, I just followed the troubleshooting guide, NOTHING! Mad
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
I know the book isn't always very useful. Did you figure out if it is the acid or bicarb that is not coming up? You may want to check the pump springs especially the acid pump. I usually can resolve these issues with the tips I've given you and have never had to do to much more than those things. Hopefully someone else may be able to help you more on this discussion board. If it is really getting to you walk away and take a break from it for a little bit.
 
Posts: 88 | Registered: 16 September 2009Reply With QuoteEdit or Delete MessageReport This Post
<Guest>
Posted
Did you check your BC volume both fresh and spent. You may have a small leak in one of the membranes, not enough to cause any other problems but a lower conductivity. Also make sure your temp is right on this can cause a difference if the condo.
 
Reply With QuoteEdit or Delete MessageReport This Post
<Guest>
Posted
Get into the habit of dumping the old calibration by pressing and holding the SVS button before you perform the new calibration. The troubleshooting manual makes a great knee pad.
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
Well, any luck??
 
Posts: 88 | Registered: 16 September 2009Reply With QuoteEdit or Delete MessageReport This Post
<Kman>
Posted
Nothing yet, i'll let you all know when I find it.
 
Reply With QuoteEdit or Delete MessageReport This Post
<bio>
Posted
Had this happen before, try a different acid pump if possible a new one, pump may be faulty even with low hours, let us know
 
Reply With QuoteEdit or Delete MessageReport This Post
<amario132>
Posted
Just wondering whats ur sodium base n bicarb base. This will determine your TCD.
 
Reply With QuoteEdit or Delete MessageReport This Post
<Dr. Machine>
Posted
13.5? Your machine is fine, don't try to fix what's not broken
 
Reply With QuoteEdit or Delete MessageReport This Post