Actuator BD NO ECHO|
Go
![]() |
New
![]() |
Find
![]() |
Notify
![]() |
Tools
![]() |
Reply
![]() |
|
I have a K model that is giving me a "actuator bd no echo" alarm. I replced the actuator board with a refurb from Fresenius. It tested well in the shop but apperared again during a treatment. Any suggestions?
|
|||
|
This is a communication error between functional board and actuator board try replacing functional board.
|
||||
|
Possible Actuator or Functional Bd problem. Can also be the deaeration motor (electrical noise). Check DeBug screen 13: If 2nd digit on the "IO" line is counting, could be motor noise. Poor Grounding can also cause this.
|
||||
|
Newesttech,
Are you talking about screen 3 ,ACT1? I don't have a screen 13. |
||||
|
I have also noticed that when I get the "no echo " alarm, my temp drops below 30.
|
||||
|
Thanks for all your help, it was the funtional board.
|
||||
|