Machine Details:

Fresenius 2008K
7K0S Series
1300 hrs.

Description of Problem:

From my observation, Machine will "spit" a tiny amount of water onto floor from the vent tube occasionally, (every 25 minutes around 1ml). The staff pointed out a huge puddle of water under the machine. I figured an internal leak, but could not find one. Still cannot locate a problem as to why the machine will spit water intermittently.


Notes:

Machine runs fine, no flow errors, no occlusions, no other leaks. We have had about 3 power failures this past week.


Thank you for your future help.


John H.

NOTE: This is my professional OPINION. You should be cautious about taking advice from a forum.
 
Posts: 19 | Registered: 19 October 2009Reply With QuoteEdit or Delete MessageReport This Post
<Techmex>
Posted
Check Incoming water regulator output psi, also a sticking float sw could be the issue.
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
Try a heat disinfect and see if the fluid amount increases. If it does very possible the heat exchanger plate has a hole in it. If so just replace the whole heat exchanger, saves time.
 