Machine Details:
Fresenius 2008K 7K0S Series 1300 hrs. Description of Problem: From my observation, Machine will "spit" a tiny amount of water onto floor from the vent tube occasionally, (every 25 minutes around 1ml). The staff pointed out a huge puddle of water under the machine. I figured an internal leak, but could not find one. Still cannot locate a problem as to why the machine will spit water intermittently. Notes: Machine runs fine, no flow errors, no occlusions, no other leaks. We have had about 3 power failures this past week. Thank you for your future help. John H. NOTE: This is my professional OPINION. You should be cautious about taking advice from a forum. |
|||
|
| <Techmex>
|
Check Incoming water regulator output psi, also a sticking float sw could be the issue.
|
||
|
Try a heat disinfect and see if the fluid amount increases. If it does very possible the heat exchanger plate has a hole in it. If so just replace the whole heat exchanger, saves time.
|
||||
|