I have rebuild acid and bicarb pumps as well as calibrated them. I have change conductivity sensor. My bicarb and acid are fresh. I have also change out spikes. Whatever I do can not make my conductivity raise any higher. I do not have flow problems or failed PHT. Thank you for all your help.
 
Posts: 127 | Location: Texas | Registered: 24 January 2001Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
We have seen this MANY times before and it has almost always been the battery RAM.
 
Posts: 875 | Location: Baltimore, MD USA | Registered: 24 October 2001Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
Where is the battery RAM located in an "h". Sorry for the stupid question.
 
Posts: 127 | Location: Texas | Registered: 24 January 2001Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
On the Functional Board. You will most likely need to change the software on the Actuator also. Before doing this, try swapping the board from another machine and see if the problem resolves. You will need to re-calibrate conductivity and blood leak when you switch the board from one machine to another though.
 
Posts: 875 | Location: Baltimore, MD USA | Registered: 24 October 2001Reply With QuoteEdit or Delete MessageReport This Post
<CP>
Posted
Remember, there are two different functional boards for the "h" machine and separate RAM for each boards. Each board has its own RAM.
 