Streamline tubing
 Login/Join
 
posted
What are the pro's / Con's of adapting your machine to this tubing ?
After your venous chambers have been calibrated can you still use combi sets if needed ?
 
Posts: 13 | Location: TAMPABAY, FLORIDA | Registered: 25 February 2011Reply With QuoteReport This Post
<calibrate>
posted
Anytime you switch lines you should recal your level detectors. The nice thing about streamline is no direct line to pressure transducer so cross contamination from that route is virtually eliminated. (the lines have a pressure diaphragm similar to old C3 cobe sets)
 
Reply With QuoteReport This Post
<Guest V>
posted
Airless system. Less need for heparin. Less clotting. Better clearances without high Qd rates.
 
Reply With QuoteReport This Post