Trying to recalibrate my TMP.
When I get to -250 cal I keep getting an operator error prompt. Any ideas why?
 
Posts: 5 | Registered: 19 September 2008Reply With QuoteEdit or Delete MessageReport This Post
<old F Guy>
Posted
If your doing it by the book and still getting the oper error it's a software glitch. Cut the machine off and reboot. It should let you calibrate then. This is true if your talking about a K anyway.
 
Reply With QuoteEdit or Delete MessageReport This Post
<Guest>
Posted
What kind of machine? Make sure your transducer connected to the meter is dry. If you don't have one there you may have gotten the inside of the meter wet. Make sure your lines are full with no air. I suppose there could be other problems but not sure what off hand.
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
K Machine, doing everything by the book, all lines dry. Will reboot.
 
Posts: 5 | Registered: 19 September 2008Reply With QuoteEdit or Delete MessageReport This Post
<breakpar now>
Posted
Same thing has happened to me. Sometimes caused by wet transducer or a kinked line. But sometimes I think it's caused by pulling the pressure to 250 too fast. I just get out of that calibration mode, get back in and go slow on drawing with the syringe and it works fine.
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
I just had this happen to me. Thought maybe it did not switch to -250 page when I hit confirm.. I just went back to options and started over. No Worries!
 
Posts: 72 | Location: Central Texas | Registered: 10 July 2007Reply With QuoteEdit or Delete MessageReport This Post
<Lady Tech #1>
Posted
WinkI have had the same problem when I was to slow in getting the -250, Reboot and start again!!! Roll Eyes
 
Reply With QuoteEdit or Delete MessageReport This Post