I have a K with 26,000 hours popping hose at the orange check vale or output to V34, already tried actuator board and cable, valves seems to be in good conditions, any more ideas are welcome
Make sure V30 is opening. If it happens when the machine is off, check your water inlet pressure regulator. I have seen the secondary side build pressure due to the heat exchange leaking from fresh to spent sides.
Posts: 124 | Location: Hawaii | Registered: 12 September 2007
if that machine has a hydroblock, with that many hours, you might want to check the rod in chamber E of the hydroblock. the diffuser plates break down and plug up the loading pressure valve as well as all of the small ports within the hydroblock...been there with those same symtoms
<Guest>
Posted
diffrod I have the same problem and it was the rod in the hydroblock. I cleaned out the hydroblock and v34 still keeps popping off. Any other places I should clean out
<bio>
Posted
Turn out to be actuator board cable, the previously change wasnt good as well, V29 and V41 check out good