I have a K machine woth a good many hours on it I had another tech working on it,He replace the whole hydroblock and everything in it,He also has changed some more things that he forgot to chart,but my problem is in heat dis. i have warter shooting out of the vent tube ,any help?
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
Most likely the heat exchanger has a pinhole leak in the thin metal plate.
 
Posts: 91 | Registered: 16 September 2009Reply With QuoteEdit or Delete MessageReport This Post
<San Augustine>
Posted
I would suggest the actuator board if you already replaced the whole hydro.
 
Reply With QuoteEdit or Delete MessageReport This Post
<Mr. Mech>
Posted
Definitely start w/ the heat x-changer. Just connect the two clear tubing together. Run heat dis., you can watch the vent tube and the open ports on the heat x-changer for leaks
 
Reply With QuoteEdit or Delete MessageReport This Post
<guess>
Posted
MR. Mech. I Didnt know the heat took part in one side of the exchanger."opeque tubing". Did i read this wrong or this what u meant
 
Reply With QuoteEdit or Delete MessageReport This Post
<botech>
Posted
Check your inlet water pressure first. If that's okay, then I'd go with the leaking heat exchanger.
 
Reply With QuoteEdit or Delete MessageReport This Post
<Mr. Mech>
Posted
Water enters thru the white tubing, as it exits thru the drain it goes thru the clear silicone tubing. Just connect the clear tubing together and run a heat diinfect, if there's no leak its the heat exchanger
 
Reply With QuoteEdit or Delete MessageReport This Post