I have a K machine woth a good many hours on it I had another tech working on it,He replace the whole hydroblock and everything in it,He also has changed some more things that he forgot to chart,but my problem is in heat dis. i have warter shooting out of the vent tube ,any help?
I would suggest the actuator board if you already replaced the whole hydro.
<Mr. Mech>
Posted
Definitely start w/ the heat x-changer. Just connect the two clear tubing together. Run heat dis., you can watch the vent tube and the open ports on the heat x-changer for leaks
<guess>
Posted
MR. Mech. I Didnt know the heat took part in one side of the exchanger."opeque tubing". Did i read this wrong or this what u meant
<botech>
Posted
Check your inlet water pressure first. If that's okay, then I'd go with the leaking heat exchanger.
<Mr. Mech>
Posted
Water enters thru the white tubing, as it exits thru the drain it goes thru the clear silicone tubing. Just connect the clear tubing together and run a heat diinfect, if there's no leak its the heat exchanger