I'm using the Dale LT544D plus for the first time on K machines. When I open the neutral I'm unable to power up the machine. Is there anyone using this tester who might have seen this?
 
Reply With QuoteEdit or Delete MessageReport This Post
<MKct>
Posted
In a 120Vac line you have Hot (black), Neutral (white), and Ground (green). Hot to neutral is your 120Vac. If you open the Neutral you will not have any power.
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
It is normal for it not to power up with the neutral open. If it does, your neutral is shorted to ground somewhere.
 
Posts: 126 | Location: Hawaii | Registered: 12 September 2007Reply With QuoteEdit or Delete MessageReport This Post
<Guest>
Posted
our biomed dept. always tests this for us. I am in the process of getting a leakage meter, can anyone elaborate on what you need/should be testing and explain it or lead me to lititure on it?
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
NFPA 99 goes into very great detail on electrical leakage and ground resistance. You'll probably have to check with your biomed department or your electricians.
 
Posts: 91 | Registered: 16 September 2009Reply With QuoteEdit or Delete MessageReport This Post