<Ron>
Posted
k machine, flow error, several different dearation pumps put in complete new one and still no pressure above 10 any help will do, have calibrated with no success
 
Reply With QuoteEdit or Delete MessageReport This Post
JP
Posted Hide Post
Valve #39 should be closed,so if you clamp the hose off there shouldn't be any change in pressure seen, but If you clamp it off and the pressure goes up then you know that valve #39 is stuck open and needs replaced. If that does not work I would try Actuator board.
 
Posts: 34 | Registered: 15 January 2008Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
Have you checked the O-ring on the restrictor in the hydroblock?
 
Posts: 205 | Location: Mountain View, CA | Registered: 19 October 2004Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
Ron, wonder what happened - solution you got from this?
I replaced the pump head - it helped a little, i encountered this long time ago
Dealer is correct too may check the o-ring
 
Posts: 221 | Registered: 16 February 2007Reply With QuoteEdit or Delete MessageReport This Post