I had a K machine that needed a Power logic Board due to a false 9V battery alarm. When I replaced the board I got this WD:Fail Long Pulse message after 15 seconds of startup. I checked the board in another machine and it works just fine. I have a feeling this is something very stupid since I'm not the original tech that troubleshot the machine.
 
Posts: 112 | Location: Arkansas, USA | Registered: 14 January 2003Reply With QuoteEdit or Delete MessageReport This Post
orc
Posted Hide Post
Check all your connections. (ie. All ribbon cables to motherboard, connection from power supply to motherboard etc.)
 
Posts: 292 | Registered: 20 January 2005Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
I just finished doing that. Earlier I found the blood pump cable loose but reconnected it with no change in the symptom.
 
Posts: 112 | Location: Arkansas, USA | Registered: 14 January 2003Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
The first time I had this problem i called the big "F". They told me it could be any one of the boards in the card cage and also it could be the power control board in the power supply. The first time it happened it was the power logic baord and the times after that it was always the actuator board.
Hope this helps, keep us posted.


Working mans PH'D
 
Posts: 45 | Location: Midwest | Registered: 30 August 2007Reply With QuoteEdit or Delete MessageReport This Post
<chieftex>
Posted
Thanks for the help guys. It was a bad actuator board.
 