beancounter,
If they are planning to use 3 mixed beds then you are 100% correct and you should find a new vendor that knows what they are doing. The third tank will serve no purpose at all other than to be a bacterial breeding ground. In instances where DI is the only treatment being used (not polishing post RO), 3 tanks are frequently used (in addition to carbon) but one is a cation, the next an anion then followed by a mixed bed. The systems such as this that I have used have had two monitors, one before the mixed bed and the other after. [This message has been edited by Chuck W (edited 10-21-2002).] |
||||
|
Theoretically, the vendor is correct .... the first tank would exhaust COMPLETELY and the second tank would exhaust nearly completely before activating the lamp indicator. HOWEVER; Chuck is also correct. The additional tank provides additional breeding ground for bacteria. The trend nationally is to get DI tanks out all together. If your RO system is functioning properly, the use of DI is redundant anyway, and a system without DI polish is far less likely to grow bacteria. I have seen MANY units, particularly hospital units, going with two ROs (either alternating or running simultaneously) if they want the protection of having back up availability. In any case, if your RO is working & maintained properly, you don't need DI on site. Many centers are leaving tank access in line & having local vendor deliver emergency tanks only if the RO goes down.
I'd be happy to answer any other questions you may have. You may reach me directly @ 502-548-0594, or you may leave a message for me @ Isopure HQ, 800-280-PURE(7873). Good Luck! |
||||
|
| Previous Topic | Next Topic | powered by eve community |
| Please Wait. Your request is being processed... |
|