When I try to calibrate the conductivity cells on a Fresenius 2000h machine I get this error: Conductivity Cells Cond less than 15 Failed, press overide calib read 15.49. What could be the problem
|
|||
|
For what reason are you calibrating (replaced cells, sensor bd, etc.)? OLC?
|
||||
|
Put the machine in a rinse mode then try to
calibrate your conductivity cells. I do not know why you changed conductivity cells. Are you receiving a conductivity stable message on the screen and what is your actual conductivity on your external meter.The H machine will not let you enter a calibration outside of any of the set parameters so that is why you are getting these errors. Have you made any other changes to the machine.Was the initial problem a flow problem or a low conductivity problem? Let us know what is going on? |
||||
|
Thanks for all of your suggestions. I found out it was the small test board. I initially had a low OLC problem. When I would try to calibrate the conductivity cells, I would get that error (however, the meter would display a cond of 13.7 give or take)
|
||||
|